Restrictions for the patient and the cardiac device system
The implanted system consists only of a lead and a pacemaker, which are each labeled
MR Conditional separately, and in combination make up an MR Conditional system.
There are no other implants in the patient’s body. For example:
Other pacemakers or ICDs
Leads no longer in use
Lead adapters
Lead extensions
The patient does not have fever.
The patient’s height is at least 1.4 meters.
The system has been implanted for at least 6 weeks.
The implanted pacing system is in the patient’s chest area.
The calculated pacing threshold does not lie above 2.0 V amplitude at 0.4 ms pulse width.
The ascertained lead impedance is between 200 and 1500 Ohms.
The implant is programmed to a special MRI mode immediately before the MRI scan.
The conditions applicable to specific devices or systems such as specific permissible
positioning zones, minimum patient height, etc. are adhered to.
Requirements of the MRI scanner
Use of a clinical MRI machine with
A closed tube
Cylindrical magnets
A static magnetic field strength of 1.5 Tesla.
The slew rate of the MRI scanner’s gradient fields should not exceed 216 T/m/s.
No additional local transmitting coils are used.
Restrictions during the MRI scan
The MRI scan can only be performed with the patient in dorsal position.
The permissible positioning zone defined below has to be adhered to.
The overall MRI scanning time accumulated from the imaging times as displayed by
the MRI scanner must not exceed 30 minutes.
The mean specific absorption rate for the whole body displayed by the MRI scanner
must not exceed 2.0 W/kg.
The head absorption rate displayed by the MRI scanner must not exceed 3.2 W/kg.
Emergency equipment for reanimation must be kept at hand and properly certified staff
must be available.
Monitor the patient’s hemodynamics during the entire MRI scan by continuously
recording at least one of the following parameters:
Blood oxygen saturation
Blood pressure
ECG
Starting from the foot end, the maximum allowed positioning mark
for the isocenter (laser light) is at the hip bone level.
Starting from the top of the skull, the maximum allowed positioning
mark for the isocenter is at the level of the eyes.
The implanted system consists only of a lead and a pacemaker, which are each labeled
MR Conditional separately, and in combination make up an MR Conditional system.
There are no other implants in the patient’s body. For example:
Other pacemakers or ICDs
Leads no longer in use
Lead adapters
Lead extensions
The patient does not have fever.
The patient’s height is at least 1.4 meters.
The system has been implanted for at least 6 weeks.
The implanted pacing system is in the patient’s chest area.
The calculated pacing threshold does not lie above 2.0 V amplitude at 0.4 ms pulse width.
The ascertained lead impedance is between 200 and 1500 Ohms.
The implant is programmed to a special MRI mode immediately before the MRI scan.
The conditions applicable to specific devices or systems such as specific permissible
positioning zones, minimum patient height, etc. are adhered to.
Requirements of the MRI scanner
Use of a clinical MRI machine with
A closed tube
Cylindrical magnets
A static magnetic field strength of 1.5 Tesla.
The slew rate of the MRI scanner’s gradient fields should not exceed 216 T/m/s.
No additional local transmitting coils are used.
Restrictions during the MRI scan
The MRI scan can only be performed with the patient in dorsal position.
The permissible positioning zone defined below has to be adhered to.
The overall MRI scanning time accumulated from the imaging times as displayed by
the MRI scanner must not exceed 30 minutes.
The mean specific absorption rate for the whole body displayed by the MRI scanner
must not exceed 2.0 W/kg.
The head absorption rate displayed by the MRI scanner must not exceed 3.2 W/kg.
Emergency equipment for reanimation must be kept at hand and properly certified staff
must be available.
Monitor the patient’s hemodynamics during the entire MRI scan by continuously
recording at least one of the following parameters:
Blood oxygen saturation
Blood pressure
ECG
Starting from the foot end, the maximum allowed positioning mark
for the isocenter (laser light) is at the hip bone level.
Starting from the top of the skull, the maximum allowed positioning
mark for the isocenter is at the level of the eyes.